41964 Patrick Wong
honolulusnow... Offline Send Email
Feb 1, 2012
2:03 pm
I took apart the failed attenuator assembly. The failure root cause was just dirty contacts. I was able to clean the contacts using electrical contact...
41965 Brooke Clarke
brookeclarke... Offline Send Email
Feb 1, 2012
5:00 pm
Hi: RMB is a great language for instrument control. AFAICR there was a correlation between the version number and the type of floppy drive you needed to load...
41966 Andrew Dobson
ez2embed Offline Send Email
Feb 1, 2012
5:23 pm
There is this link with software to a variant of 6.2: http://www.hpmuseum.net/display_item.php?sw=405 Though I have no knowledge if it is usable on anything...
41967 dpsiu Offline Send Email Feb 2, 2012
1:58 am
I have an 8720C with Opt. 006 from the factory. The solid-state transfer switch does not function any more. I replaced it with a compatible mechanical switch...
41968 Peter Gottlieb
hpnpilot219 Offline Send Email
Feb 2, 2012
3:54 am
I have been going through a checkout of my new 8566B and so far most everything is pretty close except for two things: First, the oscillator was quite high....
41969 Ed Breya
edbreya Offline Send Email
Feb 2, 2012
6:03 am
Are you sure you're looking at the fundamental frequency from the calibrator? The 8566B can sometimes be flakey when using the automatic search mode, and may...
41970 joe_kn5u Offline Send Email Feb 2, 2012
11:59 am
Patrick, Nice job keeping the group informaed - this will be invaluable information to folks who purchase one of these fine instruments that needs some work! ...
41971 Peter Gottlieb
hpnpilot219 Offline Send Email
Feb 2, 2012
12:36 pm
Yes, absolutely it is tuned to 100 MHz and at the peak. The problem is not adjustment range, it is that the level displayed (and measured) depends on where...
41972 chipbee40 Offline Send Email Feb 2, 2012
1:32 pm
Are you looking for Q7? or a board? Q7 is part number 1853-0232 I believe, these are available from various suppliers at $10-$16 a piece. JC...
41973 Brad Allen
brad50_14624 Offline Send Email
Feb 2, 2012
2:16 pm
I am actually looking for either one. I found the transistor for $10.00 each at Nebraska Surplus. The thing is, I really don't know what else might have been...
41974 w4jda Offline Send Email Feb 2, 2012
3:45 pm
I have 2 or 3 of these. Out of production for years and no other manufacturers. Used in a lot of 80's HP 8656A/B,et al. and often fail now and units parted...
41975 Ed Breya
edbreya Offline Send Email
Feb 2, 2012
4:52 pm
Normally it should be pretty good after 20 minutes from cold start. If it took hours to "settle," that could be how long it takes the internal temperature of...
41976 Thomas Miller
tmillermdems Offline Send Email
Feb 2, 2012
5:00 pm
It sure sounds like a bad oven. It might be a good idea to pull it out (from the bottom) and test it. Make sure there is power to the heater. Tom ... From: Ed...
41977 W2HX
w2hx_w2hx Offline Send Email
Feb 2, 2012
5:04 pm
Peter, can you substitute your rb reference for the internal reference? Just to see if it stabilizes faster? Or instantly? Might help determine something. 73...
41978 David Wise
vintage_dave2 Offline Send Email
Feb 2, 2012
5:20 pm
Sorry, I have 08640-90186. My change package covers 1810A through 2031A, so I guess my material is roughly equivalent to 08640-90197. As Dave H. said, it's...
41979 Bob Koller
testtech Offline Send Email
Feb 2, 2012
5:33 pm
This sounds like a log amp problem.. ________________________________ From: Peter Gottlieb <hpnpilot@...> To: hp_agilent_equipment@yahoogroups.com ...
41980 Ed Breya
edbreya Offline Send Email
Feb 2, 2012
5:43 pm
Since the amplitude display is a scaled representation of a logarithmic process, there are two adjustments needed - a gain to set the spacing of the 10 dB...
41981 chipbee40 Offline Send Email Feb 2, 2012
5:52 pm
I had the same fault on a 1741A, R26 (18.2K) and R11 (47.5K) were o/c, but no dead transistors. These scopes are plagued by bad joints on the white mulipin...
41982 Peter Gottlieb
hpnpilot219 Offline Send Email
Feb 2, 2012
6:02 pm
The unit had sat in my car at about 0 C for hours before I brought it inside. At first, the frequency moved pretty quickly but then slowed down as the...
41983 Orin Eman
orinem Offline Send Email
Feb 2, 2012
6:04 pm
... It is quite simple if the log amps are working right. Peter needs to do the log fidelity test, then go through the cal procedure. First I'd check the...
41984 Brad Allen
brad50_14624 Offline Send Email
Feb 2, 2012
9:39 pm
Thanks for the info. The board with the white square is going to stay right where it is. I am working on the Horizontal sweep board. The one transistor and the...
41985 Bruce
brucekbaur Offline Send Email
Feb 2, 2012
10:57 pm
Can't help with the wire, but if anyone needs any new rubber grommets for the attenuator, I have a number of new ones. Mine turned to rock and disinigrated. ...
41986 Peter Gottlieb
hpnpilot219 Offline Send Email
Feb 3, 2012
3:50 am
Prelude: Oven warm, was right on and rock solid after instrument in standby for a day. IF section power supplies within tolerance, no adjustment needed (well, ...
41987 Peter Gottlieb
hpnpilot219 Offline Send Email
Feb 3, 2012
4:07 am
From my verification tests, it would seem you are correct....
41988 Peter Gottlieb
hpnpilot219 Offline Send Email
Feb 3, 2012
12:30 pm
Thank you for your comment about the log amp boards being a matched set. I looked into this further and found where the manual says they are factory matched...
41989 Bob W5UQ
w5euq4 Offline Send Email
Feb 3, 2012
4:51 pm
A few months ago someone told me that there was a "kit" one could buy to refurbish these rubber membrane switches. I should have kept the email in a safe...
41990 Dan Rae
ac6ao Offline Send Email
Feb 3, 2012
5:18 pm
... Bob, you can buy a two part mixture for re-furbishing these switches. I have done it on a Fluke signal generator and several TV type remotes with success....
41991 Brad Allen
brad50_14624 Offline Send Email
Feb 3, 2012
5:29 pm
This is a normal everyday issue with telephones in the business world. The membrane has black carbon dots glued to it. There is a coating that can be placed...
41992 Orin Eman
orinem Offline Send Email
Feb 3, 2012
5:39 pm
... I'll be interested to hear what you find. Could you let us know the part numbers of the log amp boards and the values of the resistors that feed the log...
41993 hpnpilot219 Offline Send Email Feb 3, 2012
6:28 pm
Yahoo trouble posting again, it is bouncing ALL my replies, having to log in to Yahoo directly to post. My settings look fine (Why does Yahoo suck so bad?) ...
Messages 41964 - 41993 of 44566   Oldest  |  < Older  |  Newer >  |  Newest
Add to My Yahoo!      XML What's This?

Copyright © 2010 Yahoo! Inc. All rights reserved.
Privacy Policy - Terms of Service - Guidelines NEW - Help