In
MultiCal and StatistiCAL you put in an estimate for the length of the
fixture. The programs use this estimate at the lowest frequency to
resolve this phase ambiguity. This is usually not difficult if the
frequency is low enough. Then the programs use the solution at this
lowest frequency and the length estimate to estimate the phase at the
next frequency point, and so on. In this way, the programs step from
frequency to frequency, each time estimating the phase at the next
frequency point. The approach is not perfect, but it usually works in
practice.
I hope that his is helpful,
Dylan
At 09:37 AM 7/30/2008, you wrote:
Thank you Dylan
When S21=S12 there is a 180deg ambiguity in the answer because of
the
square root function. Similarly, there is a square root to solve for
S22 of the input adapter that gives another 180deg ambiguity.
The first 180deg ambiguity is ok because usually we need S12*S21
products and as long as the same 180deg error is made in calculating
both S-parameters of the input and the output adapters, the DUT
deembedding will be correct. However I'm not sure the 180deg
ambiguity
in S22 of the input adapter is as easy to deal with.
BTW, I'd like to write a matlab routine to determine the full
S-parameters of the input and the output adapters.
Regards
Igor
>
> Sorry Igor,
>
> When de-embedding adapters you can use reciprocity
(S21=S12) to
> solve this problem. Both MultiCal and StatistiCAL allow you to
solve
for
> adapter scattering parameters with multiline TRL, and that would
be
a good
> place to start.
>
> Next time I will have to be more careful about getting the
reply
> to you!
>
> Best,
>
> Dylan
>
>
> At 03:54 PM 7/22/2008, you wrote:
>
> >Hello Dylan
> >
> >Looks like you replied to my message but the text did not
show?
> >
> >Regards
> >Igor
> >
> >--- In
<
mailto:mtt-11%40yahoogroups.com>mtt-11@yahoogroups.com, Dylan
> >Williams <dylan@> wrote:
> > >
> > > At 07:46 AM 7/22/2008, you wrote:
> > >
> > > >Hello
> > > >
> > > >I've been following Doug Rytting presentation called
"TRL
> > > >Calibration". The equations all make sense and
give good results
> > > >however the presentation does not seem complete as it
does not
discuss
> > > >how to do the final step of separating e10e32 and
e23e01 terms
(these
> > > >terms are the S12's and S21's of the two error
adapters).
> > > >
> > > >Can anybody help me with this? I'd like to determine
the
S-parameters
> > > >of the two error adapters.
> > > >
> > > >Thank you
> > > >Igor
> > > >
> > > >
> > >
> > >
> > >
> > > -----------------------------------------------------
> > > Dylan Williams <dylan@>
> > > NIST 818.01 <dylan@>
> > > 325 Broadway [+1] (303)497-3138 (TEL)
> > > Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
> > >
<
http://www.boulder.nist.gov/dylan/
>
http://www.boulder.nist.gov/dylan/
> > > -----------------------------------------------------
> > >
> >
> >
>
>
>
> -----------------------------------------------------
> Dylan Williams <dylan@...>
> NIST 818.01 <dylan@...>
> 325 Broadway [+1] (303)497-3138 (TEL)
> Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
>
http://www.boulder.nist.gov/dylan/
> -----------------------------------------------------
>
-----------------------------------------------------
Dylan
Williams
<dylan@...>
NIST
818.01
<dylan@...>
325
Broadway
[+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1]
(303)497-3970 (FAX)
http://www.boulder.nist.gov/dylan/ -----------------------------------------------------
Actually knowing the type of the reflect standard corrects the S22 phase.
Igor
--- In mtt-11@yahoogroups.com, "igorlinetsky" <igor_linetsky@...> wrote:
>
> Thank you Dylan
>
> When S21=S12 there is a 180deg ambiguity in the answer because of the
> square root function. Similarly, there is a square root to solve for
> S22 of the input adapter that gives another 180deg ambiguity.
>
> The first 180deg ambiguity is ok because usually we need S12*S21
> products and as long as the same 180deg error is made in calculating
> both S-parameters of the input and the output adapters, the DUT
> deembedding will be correct. However I'm not sure the 180deg ambiguity
> in S22 of the input adapter is as easy to deal with.
>
> BTW, I'd like to write a matlab routine to determine the full
> S-parameters of the input and the output adapters.
>
> Regards
> Igor
>
> >
> > Sorry Igor,
> >
> > When de-embedding adapters you can use reciprocity
> (S21=S12) to
> > solve this problem. Both MultiCal and StatistiCAL allow you to solve
> for
> > adapter scattering parameters with multiline TRL, and that would be
> a good
> > place to start.
> >
> > Next time I will have to be more careful about getting the
> reply
> > to you!
> >
> > Best,
> >
> > Dylan
> >
> >
> > At 03:54 PM 7/22/2008, you wrote:
> >
> > >Hello Dylan
> > >
> > >Looks like you replied to my message but the text did not show?
> > >
> > >Regards
> > >Igor
> > >
> > >--- In <mailto:mtt-11%40yahoogroups.com>mtt-11@yahoogroups.com,
Dylan
> > >Williams <dylan@> wrote:
> > > >
> > > > At 07:46 AM 7/22/2008, you wrote:
> > > >
> > > > >Hello
> > > > >
> > > > >I've been following Doug Rytting presentation called "TRL
> > > > >Calibration". The equations all make sense and give good results
> > > > >however the presentation does not seem complete as it does not
> discuss
> > > > >how to do the final step of separating e10e32 and e23e01 terms
> (these
> > > > >terms are the S12's and S21's of the two error adapters).
> > > > >
> > > > >Can anybody help me with this? I'd like to determine the
> S-parameters
> > > > >of the two error adapters.
> > > > >
> > > > >Thank you
> > > > >Igor
> > > > >
> > > > >
> > > >
> > > >
> > > >
> > > > -----------------------------------------------------
> > > > Dylan Williams <dylan@>
> > > > NIST 818.01 <dylan@>
> > > > 325 Broadway [+1] (303)497-3138 (TEL)
> > > > Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
> > > >
> <http://www.boulder.nist.gov/dylan/>http://www.boulder.nist.gov/dylan/
> > > > -----------------------------------------------------
> > > >
> > >
> > >
> >
> >
> >
> > -----------------------------------------------------
> > Dylan Williams <dylan@>
> > NIST 818.01 <dylan@>
> > 325 Broadway [+1] (303)497-3138 (TEL)
> > Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
> > http://www.boulder.nist.gov/dylan/
> > -----------------------------------------------------
> >
>
Thank you Dylan
When S21=S12 there is a 180deg ambiguity in the answer because of the
square root function. Similarly, there is a square root to solve for
S22 of the input adapter that gives another 180deg ambiguity.
The first 180deg ambiguity is ok because usually we need S12*S21
products and as long as the same 180deg error is made in calculating
both S-parameters of the input and the output adapters, the DUT
deembedding will be correct. However I'm not sure the 180deg ambiguity
in S22 of the input adapter is as easy to deal with.
BTW, I'd like to write a matlab routine to determine the full
S-parameters of the input and the output adapters.
Regards
Igor
>
> Sorry Igor,
>
> When de-embedding adapters you can use reciprocity
(S21=S12) to
> solve this problem. Both MultiCal and StatistiCAL allow you to solve
for
> adapter scattering parameters with multiline TRL, and that would be
a good
> place to start.
>
> Next time I will have to be more careful about getting the
reply
> to you!
>
> Best,
>
> Dylan
>
>
> At 03:54 PM 7/22/2008, you wrote:
>
> >Hello Dylan
> >
> >Looks like you replied to my message but the text did not show?
> >
> >Regards
> >Igor
> >
> >--- In <mailto:mtt-11%40yahoogroups.com>mtt-11@yahoogroups.com, Dylan
> >Williams <dylan@> wrote:
> > >
> > > At 07:46 AM 7/22/2008, you wrote:
> > >
> > > >Hello
> > > >
> > > >I've been following Doug Rytting presentation called "TRL
> > > >Calibration". The equations all make sense and give good results
> > > >however the presentation does not seem complete as it does not
discuss
> > > >how to do the final step of separating e10e32 and e23e01 terms
(these
> > > >terms are the S12's and S21's of the two error adapters).
> > > >
> > > >Can anybody help me with this? I'd like to determine the
S-parameters
> > > >of the two error adapters.
> > > >
> > > >Thank you
> > > >Igor
> > > >
> > > >
> > >
> > >
> > >
> > > -----------------------------------------------------
> > > Dylan Williams <dylan@>
> > > NIST 818.01 <dylan@>
> > > 325 Broadway [+1] (303)497-3138 (TEL)
> > > Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
> > >
<http://www.boulder.nist.gov/dylan/>http://www.boulder.nist.gov/dylan/
> > > -----------------------------------------------------
> > >
> >
> >
>
>
>
> -----------------------------------------------------
> Dylan Williams <dylan@...>
> NIST 818.01 <dylan@...>
> 325 Broadway [+1] (303)497-3138 (TEL)
> Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
> http://www.boulder.nist.gov/dylan/
> -----------------------------------------------------
>
When
de-embedding adapters you can use reciprocity (S21=S12) to solve this
problem. Both MultiCal and StatistiCAL allow you to solve for adapter
scattering parameters with multiline TRL, and that would be a good place
to start.
Next time
I will have to be more careful about getting the reply to you!
Best,
Dylan
At 03:54 PM 7/22/2008, you wrote:
Hello Dylan
Looks like you replied to my message but the text did not show?
Regards
Igor
--- In
mtt-11@yahoogroups.com,
Dylan Williams <dylan@...> wrote:
>
> At 07:46 AM 7/22/2008, you wrote:
>
> >Hello
> >
> >I've been following Doug Rytting presentation called
"TRL
> >Calibration". The equations all make sense and give good
results
> >however the presentation does not seem complete as it does not
discuss
> >how to do the final step of separating e10e32 and e23e01 terms
(these
> >terms are the S12's and S21's of the two error adapters).
> >
> >Can anybody help me with this? I'd like to determine the
S-parameters
> >of the two error adapters.
> >
> >Thank you
> >Igor
> >
> >
>
>
>
> -----------------------------------------------------
> Dylan Williams <dylan@...>
> NIST 818.01 <dylan@...>
> 325 Broadway [+1] (303)497-3138 (TEL)
> Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
>
http://www.boulder.nist.gov/dylan/
> -----------------------------------------------------
>
-----------------------------------------------------
Dylan
Williams
<dylan@...>
NIST
818.01
<dylan@...>
325
Broadway
[+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1]
(303)497-3970 (FAX)
http://www.boulder.nist.gov/dylan/ -----------------------------------------------------
Hello Dylan
Looks like you replied to my message but the text did not show?
Regards
Igor
--- In mtt-11@yahoogroups.com, Dylan Williams <dylan@...> wrote:
>
> At 07:46 AM 7/22/2008, you wrote:
>
> >Hello
> >
> >I've been following Doug Rytting presentation called "TRL
> >Calibration". The equations all make sense and give good results
> >however the presentation does not seem complete as it does not discuss
> >how to do the final step of separating e10e32 and e23e01 terms (these
> >terms are the S12's and S21's of the two error adapters).
> >
> >Can anybody help me with this? I'd like to determine the S-parameters
> >of the two error adapters.
> >
> >Thank you
> >Igor
> >
> >
>
>
>
> -----------------------------------------------------
> Dylan Williams <dylan@...>
> NIST 818.01 <dylan@...>
> 325 Broadway [+1] (303)497-3138 (TEL)
> Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
> http://www.boulder.nist.gov/dylan/
> -----------------------------------------------------
>
I've been following Doug Rytting presentation called "TRL
Calibration". The equations all make sense and give good
results
however the presentation does not seem complete as it does not
discuss
how to do the final step of separating e10e32 and e23e01 terms
(these
terms are the S12's and S21's of the two error adapters).
Can anybody help me with this? I'd like to determine the
S-parameters
of the two error adapters.
Thank you
Igor
-----------------------------------------------------
Dylan
Williams
<dylan@...>
NIST
818.01
<dylan@...>
325
Broadway
[+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1]
(303)497-3970 (FAX)
http://www.boulder.nist.gov/dylan/ -----------------------------------------------------
Hello
I've been following Doug Rytting presentation called "TRL
Calibration". The equations all make sense and give good results
however the presentation does not seem complete as it does not discuss
how to do the final step of separating e10e32 and e23e01 terms (these
terms are the S12's and S21's of the two error adapters).
Can anybody help me with this? I'd like to determine the S-parameters
of the two error adapters.
Thank you
Igor
It still
is very difficult to measure scattering parameters with non-contact
means. Nevertheless you might find this paper our group wrote on the
subject interesting.
Are you guys talking about the
contactless probing solution? Sorry
about that if not.
A few years back, I heard that people used optoelectronic probe to do
[S] measurements without mechanical contacts? Is it getting more and
more practical and feasible?
Thanks
Fei
--- In
mtt-11@yahoogroups.com,
Dylan Williams <dylan@...> wrote:
>
> Hello Cory,
>
> If you are interested in power calibrations, I think that
Agilent,
> New Focus, Newport, U2t, perhaps other options exist. Some of these
might
> have phase calibrations as well, probably measured with an
oscilloscope.
>
> For magnitude and phase calibrations with mismatch
corrections
> traceable to electro-optic sampling, as far as I know, only NIST
and
> Anritsu have anything to offer. Our calibrations are quite
difficult and
> time consuming, so I think that Anritsu might be a better choice
for you.
>
> I hope that is helpful,
>
> Dylan
>
> At 01:01 PM 7/31/2007, you wrote:
>
>
> >Dylan,
> >
> >I contacted Anritsu, and they only support 1550nm for now, and
their list
> >price is about 25K. Do you know of any other vendors that are
selling
> >calibrated photodiodes (I think Agilent is only selling full
systems)? Can
> >you give me an order of magnitude estimate as to how much would
it
cost
> >from NIST?
> >
> >Thanks,
> >Cory
> >
> >Dylan Williams
>
><<
mailto:dylan%40boulder.ni>dylan@...
> >st.gov> To
> >Sent by:
<
mailto:mtt-11%40yahoogroups.com>mtt-11@yahoogroups.com
> >mtt-11@yahoogroup cc
> >s.com
> >Subject
> >Re: [mtt-11] What is the Most
> >07/31/2007 07:38 Practical Method to Add Optical
> >AM Capabilities to an 8510C?
> >
> >
> >Please respond to
> >mtt-11@yahoogroup
> >s.com
> >
> >
> >
> >Hello Cory,
> >
> >I believe that you are correct that the most inexpensive way
to
> >proceed is to use a calibrated photodiode. I am not sure how
much
Anritsu
> >sells their calibrated photodiodes for, but I would imagine that
would be
> >less expensive than purchasing a photodiode and calibration
directly from
> >NIST.
> >
> >Dylan
> >
> >At 07:25 PM 7/30/2007, you wrote:
> >
> >Hi,
> >
> >I am looking into what is the most practical method to add
optical
> >capabilities to my 8510C. I need to be able to measure
communication
> >laser diodes and photodiodes over the wavelength range of 850nm
to
> >1550nm (from 1Gbps to 10Gbps). I don't have the budget to
purchase
> >an off the shelf lightwave component analyzer (especially since
I
> >probably need more than one to cover this range), so I am
looking
> >into adding calibrated laser sources and detectors to my
existing
> >VNA. From what I gather the biggest challenge is to calibrate
the
> >detector, and then that can be used to calibrate the
source.
> >
> >I see many papers regarding calibrating photodetectors, but
this
> >seems to be fairly involved. Is this the only practical option,
or
> >does anyone sell reasonably priced calibrated photodetectors?
So
> >far, I have only seen a single wavelength model by Anritsu.
> >
> >For the light source, is there anything good that is
commercially
> >available, or am I best off building something myself. For
home
> >built, is it better to directly modulate or use and
external
> >modulator?
> >
> >Any feedback is appreciated.
> >
> >Thanks,
> >Cory
> >
> >-----------------------------------------------------
> >Dylan Williams
<<
mailto:dylan%40boulder.nist.gov>dylan@...>
> >NIST 818.01
<<
mailto:dylan%40ieee.org>dylan@...>
> >325 Broadway [+1] (303)497-3138 (TEL)
> >Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
>
><
http://www.boulder.nist.gov/dylan/
>
http://www.boulder.nist.gov/dylan
/
> >-----------------------------------------------------
> >
> >
>
>
>
> -----------------------------------------------------
> Dylan Williams <dylan@...>
> NIST 818.01 <dylan@...>
> 325 Broadway [+1] (303)497-3138 (TEL)
> Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
>
http://www.boulder.nist.gov/dylan/
> -----------------------------------------------------
>
-----------------------------------------------------
Dylan
Williams
<dylan@...>
NIST
818.01
<dylan@...>
325
Broadway
[+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1]
(303)497-3970 (FAX)
http://www.boulder.nist.gov/dylan/ -----------------------------------------------------
Are you guys talking about the contactless probing solution? Sorry
about that if not.
A few years back, I heard that people used optoelectronic probe to do
[S] measurements without mechanical contacts? Is it getting more and
more practical and feasible?
Thanks
Fei
--- In mtt-11@yahoogroups.com, Dylan Williams <dylan@...> wrote:
>
> Hello Cory,
>
> If you are interested in power calibrations, I think that
Agilent,
> New Focus, Newport, U2t, perhaps other options exist. Some of these
might
> have phase calibrations as well, probably measured with an
oscilloscope.
>
> For magnitude and phase calibrations with mismatch
corrections
> traceable to electro-optic sampling, as far as I know, only NIST
and
> Anritsu have anything to offer. Our calibrations are quite
difficult and
> time consuming, so I think that Anritsu might be a better choice
for you.
>
> I hope that is helpful,
>
> Dylan
>
> At 01:01 PM 7/31/2007, you wrote:
>
>
> >Dylan,
> >
> >I contacted Anritsu, and they only support 1550nm for now, and
their list
> >price is about 25K. Do you know of any other vendors that are
selling
> >calibrated photodiodes (I think Agilent is only selling full
systems)? Can
> >you give me an order of magnitude estimate as to how much would it
cost
> >from NIST?
> >
> >Thanks,
> >Cory
> >
> >Dylan Williams
> ><<mailto:dylan%40boulder.ni>dylan@...
> >st.gov> To
> >Sent by: <mailto:mtt-11%40yahoogroups.com>mtt-11@yahoogroups.com
> >mtt-11@yahoogroup cc
> >s.com
> >Subject
> >Re: [mtt-11] What is the Most
> >07/31/2007 07:38 Practical Method to Add Optical
> >AM Capabilities to an 8510C?
> >
> >
> >Please respond to
> >mtt-11@yahoogroup
> >s.com
> >
> >
> >
> >Hello Cory,
> >
> >I believe that you are correct that the most inexpensive way to
> >proceed is to use a calibrated photodiode. I am not sure how much
Anritsu
> >sells their calibrated photodiodes for, but I would imagine that
would be
> >less expensive than purchasing a photodiode and calibration
directly from
> >NIST.
> >
> >Dylan
> >
> >At 07:25 PM 7/30/2007, you wrote:
> >
> >Hi,
> >
> >I am looking into what is the most practical method to add optical
> >capabilities to my 8510C. I need to be able to measure
communication
> >laser diodes and photodiodes over the wavelength range of 850nm to
> >1550nm (from 1Gbps to 10Gbps). I don't have the budget to purchase
> >an off the shelf lightwave component analyzer (especially since I
> >probably need more than one to cover this range), so I am looking
> >into adding calibrated laser sources and detectors to my existing
> >VNA. From what I gather the biggest challenge is to calibrate the
> >detector, and then that can be used to calibrate the source.
> >
> >I see many papers regarding calibrating photodetectors, but this
> >seems to be fairly involved. Is this the only practical option, or
> >does anyone sell reasonably priced calibrated photodetectors? So
> >far, I have only seen a single wavelength model by Anritsu.
> >
> >For the light source, is there anything good that is commercially
> >available, or am I best off building something myself. For home
> >built, is it better to directly modulate or use and external
> >modulator?
> >
> >Any feedback is appreciated.
> >
> >Thanks,
> >Cory
> >
> >-----------------------------------------------------
> >Dylan Williams <<mailto:dylan%40boulder.nist.gov>dylan@...>
> >NIST 818.01 <<mailto:dylan%40ieee.org>dylan@...>
> >325 Broadway [+1] (303)497-3138 (TEL)
> >Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
>
><http://www.boulder.nist.gov/dylan/>http://www.boulder.nist.gov/dylan
/
> >-----------------------------------------------------
> >
> >
>
>
>
> -----------------------------------------------------
> Dylan Williams <dylan@...>
> NIST 818.01 <dylan@...>
> 325 Broadway [+1] (303)497-3138 (TEL)
> Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
> http://www.boulder.nist.gov/dylan/
> -----------------------------------------------------
>
If you are
interested in power calibrations, I think that Agilent, New Focus,
Newport, U2t, perhaps other options exist. Some of these might have phase
calibrations as well, probably measured with an oscilloscope.
For
magnitude and phase calibrations with mismatch corrections traceable to
electro-optic sampling, as far as I know, only NIST and Anritsu have
anything to offer. Our calibrations are quite difficult and time
consuming, so I think that Anritsu might be a better choice for
you.
I hope that is helpful,
Dylan
At 01:01 PM 7/31/2007, you wrote:
Dylan,
I contacted Anritsu, and they only support 1550nm for now, and their
list
price is about 25K. Do you know of any other vendors that are
selling
calibrated photodiodes (I think Agilent is only selling full systems)?
Can
you give me an order of magnitude estimate as to how much would it
cost
from NIST?
Thanks,
Cory
Dylan Williams
<dylan@...
st.gov> To
Sent by:
mtt-11@yahoogroups.com
mtt-11@yahoogroup cc
s.com
Subject
Re: [mtt-11] What is the Most
07/31/2007 07:38 Practical Method to Add Optical
AM Capabilities to an 8510C?
Please respond to
mtt-11@yahoogroup
s.com
Hello Cory,
I believe that you are correct that the most inexpensive way to
proceed is to use a calibrated photodiode. I am not sure how much
Anritsu
sells their calibrated photodiodes for, but I would imagine that would
be
less expensive than purchasing a photodiode and calibration directly
from
NIST.
Dylan
At 07:25 PM 7/30/2007, you wrote:
Hi,
I am looking into what is the most practical method to add optical
capabilities to my 8510C. I need to be able to measure communication
laser diodes and photodiodes over the wavelength range of 850nm to
1550nm (from 1Gbps to 10Gbps). I don't have the budget to purchase
an off the shelf lightwave component analyzer (especially since I
probably need more than one to cover this range), so I am looking
into adding calibrated laser sources and detectors to my existing
VNA. From what I gather the biggest challenge is to calibrate the
detector, and then that can be used to calibrate the source.
I see many papers regarding calibrating photodetectors, but this
seems to be fairly involved. Is this the only practical option, or
does anyone sell reasonably priced calibrated photodetectors? So
far, I have only seen a single wavelength model by Anritsu.
For the light source, is there anything good that is commercially
available, or am I best off building something myself. For home
built, is it better to directly modulate or use and external
modulator?
Any feedback is appreciated.
Thanks,
Cory
-----------------------------------------------------
Dylan Williams
<dylan@...
>
NIST 818.01
<dylan@...>
325 Broadway [+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
http://www.boulder.nist.gov/dylan/
-----------------------------------------------------
-----------------------------------------------------
Dylan
Williams
<dylan@...>
NIST
818.01
<dylan@...>
325
Broadway
[+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1]
(303)497-3970 (FAX)
http://www.boulder.nist.gov/dylan/ -----------------------------------------------------
Dylan,
I contacted Anritsu, and they only support 1550nm for now, and their list
price is about 25K. Do you know of any other vendors that are selling
calibrated photodiodes (I think Agilent is only selling full systems)? Can
you give me an order of magnitude estimate as to how much would it cost
from NIST?
Thanks,
Cory
Dylan Williams
<dylan@...
st.gov> To
Sent by: mtt-11@yahoogroups.com
mtt-11@yahoogroup cc
s.com
Subject
Re: [mtt-11] What is the Most
07/31/2007 07:38 Practical Method to Add Optical
AM Capabilities to an 8510C?
Please respond to
mtt-11@yahoogroup
s.com
Hello Cory,
I believe that you are correct that the most inexpensive way to
proceed is to use a calibrated photodiode. I am not sure how much Anritsu
sells their calibrated photodiodes for, but I would imagine that would be
less expensive than purchasing a photodiode and calibration directly from
NIST.
Dylan
At 07:25 PM 7/30/2007, you wrote:
Hi,
I am looking into what is the most practical method to add optical
capabilities to my 8510C. I need to be able to measure communication
laser diodes and photodiodes over the wavelength range of 850nm to
1550nm (from 1Gbps to 10Gbps). I don't have the budget to purchase
an off the shelf lightwave component analyzer (especially since I
probably need more than one to cover this range), so I am looking
into adding calibrated laser sources and detectors to my existing
VNA. From what I gather the biggest challenge is to calibrate the
detector, and then that can be used to calibrate the source.
I see many papers regarding calibrating photodetectors, but this
seems to be fairly involved. Is this the only practical option, or
does anyone sell reasonably priced calibrated photodetectors? So
far, I have only seen a single wavelength model by Anritsu.
For the light source, is there anything good that is commercially
available, or am I best off building something myself. For home
built, is it better to directly modulate or use and external
modulator?
Any feedback is appreciated.
Thanks,
Cory
-----------------------------------------------------
Dylan Williams <dylan@...>
NIST 818.01 <dylan@...>
325 Broadway [+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
http://www.boulder.nist.gov/dylan/
-----------------------------------------------------
I believe
that you are correct that the most inexpensive way to proceed is to use a
calibrated photodiode. I am not sure how much Anritsu sells their
calibrated photodiodes for, but I would imagine that would be less
expensive than purchasing a photodiode and calibration directly from
NIST.
Dylan
At 07:25 PM 7/30/2007, you wrote:
Hi,
I am looking into what is the most practical method to add optical
capabilities to my 8510C. I need to be able to measure communication
laser diodes and photodiodes over the wavelength range of 850nm to
1550nm (from 1Gbps to 10Gbps). I don't have the budget to purchase
an off the shelf lightwave component analyzer (especially since I
probably need more than one to cover this range), so I am looking
into adding calibrated laser sources and detectors to my existing
VNA. From what I gather the biggest challenge is to calibrate the
detector, and then that can be used to calibrate the source.
I see many papers regarding calibrating photodetectors, but this
seems to be fairly involved. Is this the only practical option, or
does anyone sell reasonably priced calibrated photodetectors? So
far, I have only seen a single wavelength model by Anritsu.
For the light source, is there anything good that is commercially
available, or am I best off building something myself. For home
built, is it better to directly modulate or use and external
modulator?
Any feedback is appreciated.
Thanks,
Cory
-----------------------------------------------------
Dylan
Williams
<dylan@...>
NIST
818.01
<dylan@...>
325
Broadway
[+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1]
(303)497-3970 (FAX)
http://www.boulder.nist.gov/dylan/ -----------------------------------------------------
Hi,
I am looking into what is the most practical method to add optical
capabilities to my 8510C. I need to be able to measure communication
laser diodes and photodiodes over the wavelength range of 850nm to
1550nm (from 1Gbps to 10Gbps). I don't have the budget to purchase
an off the shelf lightwave component analyzer (especially since I
probably need more than one to cover this range), so I am looking
into adding calibrated laser sources and detectors to my existing
VNA. From what I gather the biggest challenge is to calibrate the
detector, and then that can be used to calibrate the source.
I see many papers regarding calibrating photodetectors, but this
seems to be fairly involved. Is this the only practical option, or
does anyone sell reasonably priced calibrated photodetectors? So
far, I have only seen a single wavelength model by Anritsu.
For the light source, is there anything good that is commercially
available, or am I best off building something myself. For home
built, is it better to directly modulate or use and external
modulator?
Any feedback is appreciated.
Thanks,
Cory
In my
experience, on-wafer probes can be made to work quite nicely with TRL
standards printed on PCB. I would consider establishing your first-tier
electrical reference plane right in a microstrip line on the PCB. This
will allow you to connect your TO can directly to the PCB in the same way
you do in production, and determine its parasitics in the environment you
intend to use it in. The best way to do this generally is to use the same
transmission line in your TRL calibration as you use to connect to the TO
can in your circuit, if that is possible.
Good luck,
Dylan
At 07:42 PM 2/14/2007, you wrote:
Hi,
I am trying to measure and model a comunication laser diode (TOSA) in
a TO can. In this case it is a 1550nm wavelength DFB laser intended
for 2.5Gbps. It will be used with a dual-loop laser driver, and we
are primarily interested in the monitor photodiode response rather
than the optical response.
Since I don't have an optical VNA, I am just doing electrical
measurements from the laser diode input to the photodiode output with
a standard 2-port VNA (8510C or 8753E). The TOSA has three
terminals: laser cathode, photodiode anode, common laser anode and
photodiode cathode. I am currently using a homemade fixture composed
of two SMA PCB connectors soldered end to end, with the center pins
shortened to allow the insertion of the TOSA. I am doing a coaxial
adapter removal calibration on the VNA, and I have created additional
fixtures in open, short, and thru configurations in order to develop
a model for the fixture.
I am having a couple of issues with this approach. First, the TOSA
is normally mounted on the edge of a PCB, with the laser and common
nodes connected on top of the PCB and the photodiode node connected
to the bottom of the PCB. This is obviously a different environment
than my fixture, so there will be some discrepancy. Second, in order
to connect the common lead I either have to bend the lead to reach
the fixture or add a short piece of wire. The inductance in the
ground lead significantly affects the results of the coupling between
input and output.
I am interested in obtaining a better fixture, but I don't know if
there is something commercially availible that would work well in
this situation. My other option is to design a PCB fixture myself
(with help from our PCB group). Please let me know if you have any
suggestions or comments.
Thanks,
Cory
-----------------------------------------------------
Dylan
Williams
<dylan@...>
NIST
818.01
<dylan@...>
325
Broadway
[+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1]
(303)497-3970 (FAX)
http://www.boulder.nist.gov/dylan/ -----------------------------------------------------
Hi,
I am trying to measure and model a comunication laser diode (TOSA) in
a TO can. In this case it is a 1550nm wavelength DFB laser intended
for 2.5Gbps. It will be used with a dual-loop laser driver, and we
are primarily interested in the monitor photodiode response rather
than the optical response.
Since I don't have an optical VNA, I am just doing electrical
measurements from the laser diode input to the photodiode output with
a standard 2-port VNA (8510C or 8753E). The TOSA has three
terminals: laser cathode, photodiode anode, common laser anode and
photodiode cathode. I am currently using a homemade fixture composed
of two SMA PCB connectors soldered end to end, with the center pins
shortened to allow the insertion of the TOSA. I am doing a coaxial
adapter removal calibration on the VNA, and I have created additional
fixtures in open, short, and thru configurations in order to develop
a model for the fixture.
I am having a couple of issues with this approach. First, the TOSA
is normally mounted on the edge of a PCB, with the laser and common
nodes connected on top of the PCB and the photodiode node connected
to the bottom of the PCB. This is obviously a different environment
than my fixture, so there will be some discrepancy. Second, in order
to connect the common lead I either have to bend the lead to reach
the fixture or add a short piece of wire. The inductance in the
ground lead significantly affects the results of the coupling between
input and output.
I am interested in obtaining a better fixture, but I don't know if
there is something commercially availible that would work well in
this situation. My other option is to design a PCB fixture myself
(with help from our PCB group). Please let me know if you have any
suggestions or comments.
Thanks,
Cory
Hallo,
we are storing our measured data, e. g. s-parameters from the VNA,
transistor parameter sets, load-pull data, intermodulatioon data, etc.
in ASCII-Files, formated in Touchstone or MDIF.
We think it would be a good idea to change to XML-formats to make it
easier to import data in external software products like EXCEL and we
could write our own software faster using XML-tools for parsing.
But it makes not so much sense to define out own syntax. We woulkd
prefer to have a standard with naming and structure conventions for
microwave measurement data. Then data and application would be
portable and exchangable. We asked Agilent and R&S which might be the
ones to set standards like this, but there seems no much activity yet.
Does anybody in this group know about such activities?
Regards
Roland Gesche
Hello,
I need some help about implementation of matched
filters, or correlators, in Lab View. I am new in Lab
view and know only its basics. I need to develop
matched filters for a qpsk receiver, in Lab view. I am
receiving a signal of 600 KHz, and 1.5 volts, having
data on it.
I've Lab view 6.2 installed on my system, I'll be
thank ful if anybody guides me about which VI to use,
for matched filters, Or if there is some example
already present.
Best Regards,
/Faisal.
__________________________________
Yahoo! Music Unlimited
Access over 1 million songs. Try it free.
http://music.yahoo.com/unlimited/
Hi
This will be my first time to perform 4 port measurement with PNA.
Could you let me know what is the best way to perform 4-port
calibration with both SOLT and TRL measurement?
Thanks,
Emily
In general, we only use NISTCal for 4-port calibrations. If you are
performing 2-port calibrations, you would better off using MultiCal or
StatistiCAL. Here are some partial answers to your questions.
At 11:09 AM 2/1/2005 +0200, you wrote:
>I would like to know if you have software available to load the 12-term
>VNA calibration coefficients from statistiCAL into the HP8510 VNA.
Sorry, but I do not have such software.
>
>
>In NISTcal, I am unable to click the "DEEMBED AND SAVE 2-PORT PARAMETERS"
>button. The "devices in this directory" indicator does not switch on.
The device directory must be filled with device measurements taken by
NISTCal before this button will light up. That is a possible problem.
>I am not sure in what format file the device must be, but is currently
>named dut.s2p. In the "calibrate 2-port" menu there is another indicator
>between the thru and reflect indicators called the "Switching Terms File".
>The text-box is hidden on my PC (windows XP) and I am not sure what to
>enter there. I think that I understand the rest of the program. It would
>be nice to have a test example.
You need switch terms for a first-tier calibration, but not for a
second-tier calibration. So perhaps that is the problem. Usually we
automatically save the switch terms in a file "GXXX" where XXX is the name
of the thru measurement, but we only do that for first-tier calibrations.
>
>
>Is there a publication available on the optimum number of lines for a
>multiline trl calibration or the effect of using more lines. On page 1212
>of IEEE TRANSACTIONS ON MICROWAVE THEORY AND TECHNIQUES, VOL. 39, NO. 7,
>JULY 1991, A Multiline Method of Network Analyzer Calibration, Roger B.
>Marks, Member, IEEE it states:
>
>Although we have no proof, an
>
>optimal design seems to require one line to be a quarter
>
>wavelength long at the band center and the other line to
>
>be three times that length .
>
>Has further research been done in this field and is there a rule of thumb?
No, little research has been done. But the program FINDLEN will allow you
to design these calibration sets.
>
>
>Thank you.
>
>
>
>Yahoo! Groups Sponsor
>ADVERTISEMENT
><http://us.ard.yahoo.com/SIG=129nj9bb6/M=298184.6018725.7038619.3001176/D=group\
s/S=1705083291:HM/EXP=1107966231/A=2532114/R=2/SIG=12kostlrd/*http://clk.atdmt.c\
om/NFX/go/yhxxxnfx0020000014nfx/direct/01/&time=1107879831238949>1d8c5b09.jpg
>1d8c5b38.jpg
>
>
>----------
>Yahoo! Groups Links
> * To visit your group on the web, go to:
> *
> <http://groups.yahoo.com/group/mtt-11/>http://groups.yahoo.com/group/mtt-11/
> *
> * To unsubscribe from this group, send an email to:
> *
>
<mailto:mtt-11-unsubscribe@yahoogroups.com?subject=Unsubscribe>mtt-11-unsubscrib\
e@yahoogroups.com
>
> *
> * Your use of Yahoo! Groups is subject to the
> <http://docs.yahoo.com/info/terms/>Yahoo! Terms of Service.
-----------------------------------------------------
Dylan Williams <dylan@...>
NIST 818.01 <dylan@...>
325 Broadway [+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
http://www.boulder.nist.gov/div813/dylan/
-----------------------------------------------------
Visit the new MTT-11 Measurement Website at www.mtt.org/measurements for
answers to all your microwave measurement questions. You will find oodles
of useful measurement links and information, as well as nine friendly
measurement forums moderated by MTT's top measurement experts !
I would like to know if you
have software available to load the 12-term VNA calibration coefficients from
statistiCAL into the HP8510 VNA.
In NISTcal, I am unable to click the "DEEMBED
AND SAVE 2-PORT PARAMETERS" button. The "devices in this
directory" indicator does not switch on. I am not sure in what format file
the device must be, but is currently named dut.s2p. In the "calibrate
2-port" menu there is another indicator between the thru and reflect
indicators called the "Switching Terms File". The text-box is hidden
on my PC (windows XP) and I am not sure what to enter there. I think that I
understand the rest of the program. It would be nice to have a test example.
Is there a publication
available on the optimum number of lines for a multiline trl calibration or the
effect of using more lines. On page 1212 of “IEEE TRANSACTIONS ON
MICROWAVE THEORY AND TECHNIQUES, VOL. 39, NO. 7, JULY 1991, A Multiline Method
of Network Analyzer Calibration, Roger B. Marks, Member, IEEE” it states:
“Although we have no
proof, an
optimal design seems to
require one line to be a quarter
wavelength long at the band
center and the other line to
be three times that length”.
Has further research been
done in this field and is there a rule of thumb?
hi all,
I am a masters student at Indian institute of
science, bangalore and am working on RF MEMS based
distributed phase-shifter.
I had a doubt regarding the modelling of the
substrate losses in a periodically loaded transmission
line.
Is there any possible way to find an equivalent
circuit for the substrate losses, probably some
resistance(depending on the measured attenuation
values) can be incorporated in the equivalent circuit.
Please clarify.
Also do you have any information regarding the effect
of DC bias on substrate losses i.e. when we apply DC
to the central conductor in CPW,for actuation, does it
effects the substrate loss .
Thanks
Indrajit
__________________________________
Do you Yahoo!?
Find out what made the Top Yahoo! Searches of 2003 http://search.yahoo.com/top2003
------------------------ Yahoo! Groups Sponsor
--------------------~-->
Make a clean sweep of pop-up ads. Yahoo! Companion Toolbar.
Now with Pop-Up Blocker. Get it for free! http://us.click.yahoo.com/L5YrjA/eSIIAA/yQLSAA/bGYolB/TM
--------------------------------------------------------------------~->
hi all,
I am a masters student at Indian institute of
science, bangalore and am working on RF MEMS based
distributed phase-shifter.
I had a doubt regarding the modelling of the
substrate losses in a periodically loaded transmission
line.
Is there any possible way to find an equivalent
circuit for the substrate losses, probably some
resistance(depending on the measured attenuation
values) can be incorporated in the equivalent circuit.
Please clarify.
Also do you have any information regarding the effect
of DC bias on substrate losses i.e. when we apply DC
to the central conductor in CPW,for actuation, does it
effects the substrate loss .
Thanks
Indrajit
__________________________________
Do you Yahoo!?
Find out what made the Top Yahoo! Searches of 2003
http://search.yahoo.com/top2003
Hi Dylan,
It's been a while...!, I tried to use the macro that you've sent to me
that makes the transformation from S4p to your program format, unfortunatly
I was not successful. Do you want me to send you the measured files in S4p?
I tried to loggin the MTT-11 web site but I do have a problem right now, do
you know how to recover the ID and the password ?
Regards
Ali
-----Original Message-----
From: Dylan Williams [mailto:dylan@...]
Sent: Monday, February 04, 2002 11:14 AM
To: mtt-11@yahoogroups.com
Subject: Re: [mtt-11] Duplicate?
Hi Roger- Jeff was working on setting up the question and answer "clubs" we
discussed earlier. The first step was understanding how the yahoo sites
work, so he asked us to be his guinea pigs while he figured out what he
could do within the yahoo structure. Jeff is now working on a sample web
page for the clubs that will allow IEEE members to do things like sign up,
submit a question, browse responses, and look through a bibliography. He
was not trying to recreate our internal MTT-11 discussion group. Dylan
At 08:57 AM 2/4/2002 +0000, you wrote:
>Hi all:
>
>We seem to have created a duplicate group since mtt-11 already
>exists!!!!!!
>
>Regards,
>
>Roger.
>
>
>
>
>
>
>To unsubscribe from this group, send an email to:
>mtt-11-unsubscribe@yahoogroups.com
>
>
>
>Your use of Yahoo! Groups is subject to http://docs.yahoo.com/info/terms/
-----------------------------------------------------
Dylan Williams <dylan@...>
NIST 813.01 <dylan@...>
325 Broadway [+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
http://www.boulder.nist.gov/div813/dylan/
-----------------------------------------------------
To unsubscribe from this group, send an email to:
mtt-11-unsubscribe@yahoogroups.com
Your use of Yahoo! Groups is subject to http://docs.yahoo.com/info/terms/
Hi Roger- Jeff was working on setting up the question and answer "clubs" we
discussed earlier. The first step was understanding how the yahoo sites
work, so he asked us to be his guinea pigs while he figured out what he
could do within the yahoo structure. Jeff is now working on a sample web
page for the clubs that will allow IEEE members to do things like sign up,
submit a question, browse responses, and look through a bibliography. He
was not trying to recreate our internal MTT-11 discussion group. Dylan
At 08:57 AM 2/4/2002 +0000, you wrote:
>Hi all:
>
>We seem to have created a duplicate group since mtt-11 already
>exists!!!!!!
>
>Regards,
>
>Roger.
>
>
>
>
>
>
>To unsubscribe from this group, send an email to:
>mtt-11-unsubscribe@yahoogroups.com
>
>
>
>Your use of Yahoo! Groups is subject to http://docs.yahoo.com/info/terms/
-----------------------------------------------------
Dylan Williams <dylan@...>
NIST 813.01 <dylan@...>
325 Broadway [+1] (303)497-3138 (TEL)
Boulder CO 80305 USA [+1] (303)497-3970 (FAX)
http://www.boulder.nist.gov/div813/dylan/
-----------------------------------------------------
Dear Newsgroup Members,
I thought I'd try sending the first message as a test since we have 9
people signed up so far. Feel free to post to the newsgroup with any
comments or questions. After we're convinced that this type of newsgroup
works satisfactorily or we make modifications, I will begin working with
Dylan Williams in starting our first specific Measurement Club, dealing
with on-wafer measurements. It will be a bit more involved than the general
newsgroup, so I figured I'd start with this general one first.
Best regards,
Jeff